V4 noise floor calibration setup

4 Posts
3 Users
0 Likes
9,015 Views
 JFTN
(@jftn)
New Member
Joined: 10 years ago
Posts: 3
Topic starter  

About the calibration level noise floor with a load of 50 ohms.
Spectran Analyzer v4.
Internal amplifier should it be activated for this operation to take into account the noise factor?
Regards


   
Quote
(@moderator)
Noble Member Admin
Joined: 14 years ago
Posts: 1044
 

Yes, allways calibrate the complete setup you want to use afterwards.

Snip


   
ReplyQuote
(@joaoo)
New Member
Joined: 10 years ago
Posts: 1
 

When I use the function Callrun with the calibration resistor connected the signal disappear and I have to restart the program and disconnect/reconnect the HF-60105 V4X to see the signal again.
Why it happens?


   
ReplyQuote
(@moderator)
Noble Member Admin
Joined: 14 years ago
Posts: 1044
 

Thats normal because the new data need to get reloaded after a system restart.

Snip


   
ReplyQuote